Tue Apr 24 2012

Sample Flow Rate

Does anybody know the sample flow rate of the Philips M1026B?  For some reason I thought it was 120mL/min but I cannot actually find any documentation to back up that number.  Thanks.


   RE: Medical Gas Monitor Report Post
Tue Apr 24 2012Reply from MedWrenchManager
MedWrenchManager

The flow on the M1019A is 200 ml/min. I think the "bench" is the same. Good luck!! 

Answer provided by Sean W., Biomedical Equipment Specialist at Thomas Jefferson University Hospital


Report Post
Wed Apr 25 2012Reply from M. Shelton
avatar placemark

The M1026A model is 150ml in normal mode and 310 ml in purge mode.  I believe the "B" model is the same.


Report Post
Mon Oct 07 2013Reply from AGM-Parts
avatar placemark

Correct, make sure you have the calibrated water trap in the front of it too.


Report Post