Refer to operator’s manual for detailed instructions.
Weekly
1. Clean external and internal surfaces
2. Change water in reservoir with clean, distilled water.
Monthly
1. Flush system
-Drain and fill reservoir with clean distilled water.
-Add one ounce of SPEED-CLEAN to cool chamber.
-Run one cycle at 270?F for six minutes. Instruments should not be sterilized during cleaning.
-Drain reservoir and refill with clean distilled water.
-Run one unwrapped cycle.
-Drain reservoir and after cooling clean trays, tray plate, rack, and filter.(refer to operators manual for details)
-Refill with clean, distilled water.
2.Check pressure relief valve.