I need to know the RESET procedure to get rid of the "service wrench" indicator on the units.After testing with Defib analyzer the indicator remains on,do you need special software to flash the unit ?Can anyone help with this.Thankyou.
Wed Mar 15 2017
To Reset AED lifepack 500
Wed Mar 15 2017Reply from MedWrench MM
If you are in need of immediate assistance please feel free to reach out directly to our Defibrillator sponsors. They are happy to help out when they can.
Just be sure to tell them you were referred by Jonathan at MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Thu Mar 16 2017Reply from Cipher
The service manual has the procedure for reviewing and clearing the fault codes. I'll send in to MedWrench Admin for posting.
- Cipher
Thu Mar 16 2017Reply from MedWrench MM
You can now find the Service Manual listed on the product page under the documents/manuals section. Hope this helps.
Jonathan F. Payne,
866-989-7057,
Thu Mar 16 2017Reply from Mariusmul