Wed Apr 25 2012

Power Supply

Is there any company / vendor that can repair the 9300xsp power supply?


   RE: Light Source Report Post
Thu Apr 26 2012Reply from MedWrenchManager
MedWrenchManager

OK, before you send it out for repair, check the 67F090 Thermostat. It is on the heat sync. I have found a few very open. That is no meter needed to see that it was open. Good Luck! 

Answer provided by Sean W., Biomedical Equipment Specialist at Thomas Jefferson University Hospital


Report Post
Wed May 02 2012Reply from MURALI
avatar placemark

Thank You! Murali