MedWrench is available to help people with specific questions and troubleshooting problems. Could you please be more specific with the issue that you are having with your instrument , and the MedWrench team will do their best to assist you in solving the issue. This will also help other community members who face similar challenges.
Jayme McKelveyVice President of Sales | MD PublishingPhone: (800) 906-3373Follow me on Facebook or LinkedIn
Fri May 04 2012Reply from GaryBiomed
Fri May 04 2012Reply from Redhead1
avatar placemark
Thank you
Thu Jan 23 2020Reply from Shamjit
Shamjit
LTV 1000 , it is not working in internal battery . Replaced the internal battery and charged for 8 hours .Still charge status LED is Red only.
Kindly advice how to sort out this issue. Internsl battery is alright
Thu Jan 23 2020Reply from jroberts
jroberts
If the internal battery isn't powering the device, replace the power board.  This board should be replaced every two years during the PM anyways, and requires pretty extensive testing after install.   Might be wise to farm out to a vendor!
-J
Fri Jan 24 2020Reply from Shamjit
Shamjit
Hello sir
Thank you so much for your advice. Can you send me the vendor details (e-mail id) of spare parts supplier especially power board of LTV 1000 , also PM kit   .Thanks in advance