Mon May 22 2017

Stop spinning

Device intermittently stops spinning with 180s flashing, then 30s will flash on subsequent attempts.

   RE: Centrifuge Report Post
Tue May 23 2017Reply from Bob S
avatar placemark

30 flashing = motor failed to reach rpm within 30 seconds

120s flashing = cover opened or cycle changed during operation

180s flashing = insufficient power to maintain RPM

30s and 180s flashing = rotor spun backward direction during reverse braking

Error/Service light will continuously illuminate when the centrifuge has achieved a total cycle count of 18,000, which

is the useful life of the drive system. Drive mechanism needs replacing. Contact an authorized service center.


Report Post
Fri Jun 30 2017Reply from Bob S
avatar placemark

180 Flashing = Insufficient Power to maintain RPM

120 Flashing = cover opened or cycle changed during operation

30 Flashing = Rotor failed to reach RPM within 30 seconds

30 + 180 Flashing = Rotor spun direction during reverese braking.


Report Post