Settings are off when the EKG prints out, the leads are not listed in the right order.
Mon May 22 2017
Settings are off when the EKG prints out, the leads are not listed in the right order.
Mon May 22 2017Reply from mcooper
Tue May 23 2017Reply from MedWrench MM
Hi Mcooper,
If you are in need of immediate assistance please feel free to reach out directly to our Electrocardiograph (EKG/ECG) sponsors. They may be able to help.
Just be sure to tell them you were referred by MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Wed May 24 2017Reply from Cipher
I'm sending a copy of the M1770A service manual to MedWrench for posting. The Operations section of the manual shows the options for lead configuration. You may need to drill into the Global COnfiguration settings as weel to set up the desired lead order. Hope this helps. - Cipher
Report Post
Wed May 24 2017Reply from MedWrench MM
Hi MCooper,
I have added numerous manuals to the product page. You will find them all listed under the manuals/documents section.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post