Good afternoon. Siemens Arcadis Varic when photographing in the picture there is nothing, and the system gives an error 5007 generator. Tell me, please, where to look for a malfunction.
Wed May 24 2017
error 5007
Wed May 24 2017Reply from MedWrench MM
If you are in need of immediate assistance please feel free to reach out directly to our Mobile C-Arm sponsors. They may be able to assist you.
Just be sure to tell them you were referred by MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Sun Sep 17 2017Reply from Tomelgy
5007 is max tube filament current exeeded. This could be D2 PCB, D1 CPU PCB or Tube.
Most likely your tube is damaged.
Set 'manual' kV (40kv & stop on), press & hold radiation button. Increase kV to see at which point the error will appear. If happens at high kV again it could be tube arcing.