Hello,
I need to calibrated the downstream occlusion on a sigma spectrum pump. Would anyone know what the code is, if there is one. Much appreciated. Thanks.
Fri May 26 2017
calibration code
Fri Sep 08 2017Reply from MedWrenchManager
If you are in need of immediate assistance please feel free to reach out directly to our Infusion Pump sponsors. They are always happy to help out when they can.
Just be sure to tell them you were referred by MedWrench.
Report Post
Fri Sep 08 2017Reply from ron.leinbach_9519
you cannot calibrate the sensor, you must send it back to the factory for calibration. I would suggest fist cleaning off the pressure plate by points 3 and 4 first, as sometimes it can get residue on it.