Is there any way to set the value myself instead of letting the unit establish the value itself for the fluid side occlusion?
Fri Jun 02 2017
Fluid Side Occlusion Calibration
Tue Jun 06 2017Reply from Cipher
Fluid side occlusion is adjustable between 1 to 15 psi for each mode, however; this can only configured using the FMS software via configuration option #28-Occlusion Alarm Incremental. The FMS kit can be purchased with or without interface cables. Hope this helps. - Cipher