I just replaced the nvram on the cardiocap 5. Now, the monitor is not displaying data from the NESTPR. The menus are on the screen, but no data. I can't start a bp. The spo2 probe lights up, but does not display data. Any thoughts?
Wed Jun 21 2017
Cardiocap 5
Thu Jun 22 2017Reply from MedWrench MM
If you are in need of immediate assistance our Telemetry-Central Station sponsors may be able to help.
Southwestern Biomedical Electronics
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Thu Jun 22 2017Reply from lakeside wholesale medical
Thu Jun 22 2017Reply from MedRepair Rx, LLC
Did you go into service mode and reset the options? When you replace the NvRam you must do this to give the monitor back the options it had or nothing will show up.
MedRepair Rx, LLC
Office: 864-757-8799
Fax: 864-757-8798
www.MedRepairRx.com
info@medrepairrx.com
Thu Jun 22 2017Reply from Ray Brown, CBET
My SWAG would be, as complicated as it is to disassemble and reassemble the CardioCap 5, I'd say you missed a cable connection inside. The menu is present, so I'd say that the option is turned on, but with the missing connection, there's no data coming in.