Error 2800 "HUB"

receiving error HUB 2800, states communication error, I can unplug, error goes away but returns about 20 minutes late while injector is just sitting there not being used. Suggestions??

   RE: Contrast Media-Injector System Report Post
Fri Jun 23 2017Reply from Crash
avatar placemark

Inside the base is a network hub used for subsytem communications. You could have a loose connection or a bad hub. Newer injectors or Certegra Workstation systems will have a switch instead of a hub. Reseat all connections. If that does not work, you may need to replace the hub/switch. 


Report Post
Fri Jun 23 2017Reply from Ryan C.
avatar placemark

I agree with the preivous about the hub inside the poewr box.  If you have a second system, you could swap the poewr boxes between the two units to confirm.  From there, I would chekc cabling and consider replacement of the hub.

RF Clarke
Injector Support and Service
888-667-1062