The ECG machine is printing all leads except the V leads. How do I solve this problem?
Wed Aug 09 2017
Printing
Wed Aug 09 2017Reply from Alexandria
Wed Aug 09 2017Reply from Jaiyeola O
Wed Aug 09 2017Reply from Chris-CBET
is there a blank line on the readout(possible print head issue) is the V5 coming out flat(check for broken cable, poor connection)
Report Post
Wed Aug 09 2017Reply from TSS Jim
Does it say Overload next to the V leads. Is there anything on the paper besides the 6 tracings from the limb leads. This might be a setting that got changed.
Report Post
Wed Aug 09 2017Reply from Jaiyeola O
TSS Jim- It prints the limb leads page only completely omitting the second page which has the v leads data. I was thinking it was settings issue as well but I couldn't find anything in set up that was changed.
Report Post
Wed Aug 09 2017Reply from Jaiyeola O
Thu Aug 10 2017Reply from TSS Jim
There is no setting to just print limb leads. I suggest going into Setup - ECG and choose a different format like 3+3.