Sun Aug 20 2017

Error

How I can fix the Vacuum Low in mindray bc-3200

   RE: Hematology Analyzer Report Post
Mon Aug 21 2017Reply from Mubark
avatar placemark

My Device its not complete the self test to work normal, because the VACUUM LOW error is come on, I followed the service manual but still I have the same error massege, Please I need support here.


Report Post
Wed Sep 02 2020Reply from cliangue
avatar placemark

hello i'm dr. Manuel . when you have vacum error. 1. You must go to the menu and ask about the status of your vacum if the pressure is greater than 12.0 is good, you get it by doing several tests until you reach it. Otherwise, disconnect the tubes from the same vacum pump. If it makes you puff, the pump is fine, otherwise, change the pump. Inside the machine there is a tube that does not transport water, it is a pressure tube, you must adjust it well like the other vacum tubes. Thank you