Service watchdog
RE: Pulse Oximeter Report Post
Were you able to figure it out?
we had a Rad7 power down shortly after it was powered up.
the unit would beep 10 times.
according to the service manual (Radical-7 Signal Extraction Pulse CO-Oximeter Service Manual 6-3) and a Masimo Tech, this was a "Docking Station Watchdog Failure".
Ours was sent in for repair....
Report Post
I was not able to find the problem so we sent this one out to Masimo. They won't sell us parts and they have been very problematic, so we're always sending them out. Thanks for your help
Derek Snyder, CBET
Report Post
There is not much that you would want to change on those. I am not sure if cases could be ordered but batteries can be. Other than that, we send them out as well. The flat rate repair makes it easy and simple.
Report Post
There was a recall issue with the software versions on the 7's.. Did you get yours updated? Did you get ones which were already updated? We bought some which were not updated and had to update the software and resolved the issues we had.
Might be a good idea or place to start. I have the recall notice, but can not load it in here for some reason.. Its a PDF..
mrstan.
Report Post
Mine are showing "error in data storage" and shuts down. Does that ring a bell with anyone out there?
mrstan
mrstan.