I'm having a problem with the boiler flashing a 0 during the cycle . I hit reset and it continues the cycle. I've changed all the exhaust manifold S3,S40,S37.
Can anyone help me out the boiler is s Reimers Electra steam
Thanks
Fri Sep 08 2017
Boiler keeps flashing 0
Fri Sep 08 2017Reply from Cbet_Matt
I don't have much experience with these it here my thoughts. Have you verified your steam supply pressure with a gauge? Does the boiler have a pressure transducer to signal the PLC? Call Steris tech support with boiler mod# and sn#. 800-333-8848.
Matt
Report Post
Wed Sep 20 2017Reply from barkyboy
try cleaning and replacing the water level sensing probes. Flush the boiler and see if that helps to correct the problem.
Barkyboy
Report Post
Mon Sep 25 2017Reply from Kara Crissman
Dear Dmplumb,
We came across your post about the Century V120 Sterilizer and we would like to help troubleshoot with you. Please provide us with your preferred method of contact or call our Service Support team at 1-800-333-8828.
Thank you,
Kara Crissman
STERIS Corporation