Wed May 23 2012

Philips optmus 50

My optmus 50 does not expose. when i expose the control panal does not display any error message. can any one help me by sending service manual.

Thnx

 

Irshad


   RE: R/F Systems Report Post
Sun Mar 24 2013Reply from eb66401
avatar placemark

I am having exactly the same problem, I hear the high speed rotor during prep and then pressing shoot, nothing. No error response. The ready light and high voltage indicators are showing green on the optimus console.


Report Post
Tue Mar 26 2013Reply from www.PI.solutions
www.PI.solutions

When you press test on the overhead does it give a message?



Jamison Ashby, MBA

Owner: Precision Imaging solutions

www.PI.solutions
Report Post
Tue Mar 26 2013Reply from eb66401
avatar placemark

I will check that tomorrow. I also will be looking close at the limit switch for the cartridge since exposure will take place for above table exposures, I recently discovered. Thank you.


Report Post
Sat Mar 30 2013Reply from Serdar
avatar placemark

it seems rotor controller is defective. But firstly you must take error logs from optimus generator. OR you can check u-v-w check points behind of optimus. it must be about 10 ohms.

FIRSTLY you must take error logs.


Report Post
Tue Apr 02 2013Reply from eb66401
avatar placemark

Thank you for all the replies and good ideas. I have found that the limit switch for the cartridge in both the table and vertical Bucky is bad. I've repaired the Switches and all is fine. Thanks again.


Report Post
Mon Aug 11 2014Reply from Abdoxman
avatar placemark

i have the same but i can`t get error log cause i don`t have the service soft , can anyone help please


Report Post