Charging Issue
RE: Mobile Radiography Report Post
You stated you performed the CAL VOLT METER procedure, but you didin't say if you performed the CAL GENERATOR procedure right after the for the volt meter calibration? This is critical, especially if you had a +/- 0.2V differerence between the voltage display when compared to actual battery voltage measurement during the CAL VOLT METER procedure. - Cipher
Report Post
Report Post
Yes, sorry, we also did the Cal Generator calibration, still giving the same issue. We will have to try the LOOP TEST tool as you described. Is it possible to email me that procedure?
Report Post
Okay, as I was driving down the highway I was monitoring replys to your situation. I happened to be talking with Master Yoda, a 40 year GE field engineer and support guy. We have two thoughts for you...Cipher may be on to something. If you load defaults it will reset the battery metering routines which will then require a full re-calibration.
However, before you do that make sure you don't a) have a bad battery and/or b) there is not a poor connection with your new battery pack.
You should have a battery test strip. Make sure each batter is within 0.2 volts of all the others.
I don't know where you bought your batteries but Yoda and I agree that Holden Battery service is as good as they come.
Good luck,
Report Post
I have had that problem many times. I'll give you the procedure to correct that.
1. Begin with service switch in run mode (up position) and power on machine.
2. The boot sequence will began, after 3 second flip service switch down for diagnostics mode
3. Use KVP buttons to scroll the option until you see diagnostics mode the press MAS down to enter.
4. The system will prompt you for pass word which is MAS up MAS down KVP down KVP up (in the formation of a backwards U).
5. Continue by using KVP buttons to scroll to the system diagnostics option and press MAS down to enter.
6. Use KVP buttons to scroll to loop test and press mass down to accept. I will ask you for an address.
7. The address is ocb4 then press mas down. (Enter numeric values by pressing KVP UP and Alphabetical values by press KVP down.)
8. Press Mas up for HEX
9. Press KVP down for write and press mas down to accept
10. Then you will need to type FF and mas down to begin the looping process
11. Once you see looping on display hit mass down3 times and us KVP to scroll to end service and follow instructions.
Report Post
This morning, I sent the step-by-step document that Catfish spelled out in his reply, to the MedWrench Admin. I was informed that they are running behind on posting docs. The backwards U mentioned in the step 4, can be confusing to some. In order, the backwards U button press is: MAS up arrow, MAS down arrow, KV down arrow, KV up arrow. - Cipher
Report Post