safety filter

looking for an assembly to replace the physician safety filter, error 23 conditions.

   RE: Surgical Laser Report Post
Wed Oct 18 2017Reply from MedWrenchMia
MedWrenchMia

Hello! What product are you referring to? Make/model.



Mia DeBiase Sales & Marketing Assistant mia@medwrench.com

Report Post
Wed Oct 18 2017Reply from gspooner
avatar placemark

Hello!  Strange that the make/model info didn't make it into my initial comment!

It's a Zeiss VisuLase 532.  Serial # 844371 (from 2002).  I have not seen this unit.  It is in Chittagong, Bangladesh, where I'll be doing some training next month.  Apparently it is installed with a Zeiss slit lamp (not sure of model).

Anyway - thanks very much!

Best,

Greg


Report Post
Tue May 29 2018Reply from Edgar23
avatar placemark

service manual visulas 532s, please in ed230384@gmail.com


Report Post
Wed Oct 14 2020Reply from Alessio
avatar placemark

hello,

can anyone tell me the procedure to be performed by mistake 23?

does the operator filter need to be replaced or just clean it? can you have the service manual?

Thx


Report Post
Wed Oct 14 2020Reply from Ben C
Ben C

Have you checked the manual


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Wed Oct 14 2020Reply from Wayne-UML
Wayne-UML

You correct in that it is the Physician's safety filter. Specifically that the open-close time is too long (>50ms). If the filter has not been used in a while, then the first step should be to clean the mechanical parts. After that, check the cable and any electrical components in the device.


Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Report Post