If the timer is not increasing or decreasing time but it can start and stop centrifuge,what could be wrong?
Wed May 30 2012
Timer
Thu May 31 2012Reply from Jayme MDP McKelvey
Posted on behalf of Byron P., Biomedical Specialist at ARAMARK Healthcare...
"It sounds like the timer and speed controls are locked. If you try to adjust either speed or time you get a "Lo on" indication.
To remove the lock, press and hold the OPEN and SHORT buttons simultaneously for 5 sec's with the lid open. The display should indicate "Lo OF", adjustments can then be made.
The opposite is true, if you hold the OPEN and SHORT buttons for 5 sec's witht the lid open, it tuns on the Lock, "Lo on". "