Our cardiocap 5 co2 readout is reading in percent. How can it be changed back to partial pressure?


   RE: Telemetry-Central Station Report Post
Fri Jun 01 2012Reply from Jean Monteiro
avatar placemark

Enter o the MAIN MENU and select INSTALL/SERVICE.

The monitor will ask you a password. the numbers are 16 - 4 -34

After you put the password you have to select the option INSTALL.

after this, go on trought the option PARAMETERS and after this you click on GASES

 

then you will have 3 options
Kpa - % - mmHg

 

Set the best choice

 

Best regards

 

Jean


Report Post
Mon Jul 01 2013Reply from Steron
avatar placemark

Hi, i hope you help me. do you know how to change de digital and waveforms area and save the changes?


Report Post
Tue Jul 02 2013Reply from Rix
avatar placemark

If I recall correctly, in order to retain those values as default (on power up) you must select "save" ( I forgot which menu) and then use the code 13-30-31.


Report Post
Tue Jul 02 2013Reply from Steron
avatar placemark

Thank you so much.  i´ll check with the password. the option Save mode you can change the waveforms and digital area, when i turn off the monitor and turn it on again doesn´t save the changes. So, i will try with the password. Thanks again.


Report Post
Sun Jul 07 2013Reply from Steron
avatar placemark

Hi Rix. Bad news.. I tried with 13-30-31 but it´s not correct. There was not changes when i turned on cardiocap. do you have another code ? By the way, thanks so much for you support. 


Report Post
Mon Jul 08 2013Reply from barkyboy
avatar placemark

try 26-23-8



Barkyboy
Report Post
Mon Jul 08 2013Reply from renglm
avatar placemark

This is older, answer was helpful for me:

A new post has been made to the MedWrench Datex Ohmeda Cardiocap/5 Questions and Answers forum.

Posted By: aneserv medical Inc.

Subject: No waveforms

Message:

First code is 26-23-8 then enter service 16-4-34

To view the posts in this forum topic, please click here.

http://www.medwrench.com?community.posts/threadNo/2941

 


Report Post
Mon Jul 08 2013Reply from renglm
avatar placemark

Sorry, after rereading, I'm not sure:

Here you find the manuals, the codes should be the technical reference.


Report Post
Tue Jul 09 2013Reply from Aneserv Medical Inc.
Aneserv Medical Inc.

Those are the correct numbers to enter the menu to save mmhg but if the monitor forgot mmhg and went to % then the CPU battery is dead...so after you change that battery you may have to reconfigure the monitor...depending on it's age if it will remember it's configuration. If you have to configure you will need the other save #...which I have but not right here at the moment.


Report Post
Wed Aug 14 2013Reply from Steron
avatar placemark

Hi again. I´ve been busy sorry. I tried with the service password, but i can´t save the changes. So, someone knows mor codes to try ? I need to save the digital and waveforms changes.

 