long cycle times
RE: Sterilizers/Autoclaves Report Post

Auxo Medical provides medical capital equipment and services. For more information see our site at www.auxomedical.com
Report Post

I figured it out, the steam trap was pluged, and the pressure regulator was adjusted down farther than it should have been.
with cleaning the steam trap, and proper adjustment of the pressure regulator unit ran an average of 7 seconds over projected cycle times.
Report Post

Thank you Jay and Mia for the suggestions, I will keep those in mind if i have any other issues.
Report Post

Have you verified that S2 & S9 are opening fully and the seat for each valve is clean? Or
does the pressure drop below 40PSI during conditioning? If the pressure drops that low you either have a bad heating element or lost power to one elements. Test the voltage at the input of the elements contactor and measure the continuity of the elements “with power off”. Heating elements should be ~ 22-28 Ohms.
Matt
Report Post

@Cbet_Matt
Are you refering to the optional steam generator?
My unit does not have the steam generator, it operates on facility steam.
Report Post

If you steam trap was plugged with debris you might want to check if there is a screen in the chamber drain.
Matt
Report Post

I checked it.
The Steam trap I was refering to was external to the machine.
Sorry for the confusion.
Report Post

I assume when you say you've checked the steam trap, you mean the trap that is part of the manifold under the chamber? Not the the stand alone jacket trap. Also, Steris gauges suck. You need to know an accurate chamber and jacket pressure to determine correct function. Always keep a couple good gauges on hand. Could be as simple as a steam regulator adjustment.
Report Post

I checked that one also, it was ok.
I agree with the Guages, i have a a couple pressure and vauum guages i bought, decided to do so when the steris guage said it was at 40 psi in the door seal when i had it removed to replace it...lol