long cycle times
RE: Sterilizers/Autoclaves Report Post
Heat exchanger and check valves, drain valve and trap of chamber should help
Mark S Fisher - Vice President
Contract Sales & Service Manager, Auxo Medical LLC
www.auxomedical.com
Report Post
Sounds like the boiler is not producing enough steam.
First I would check that the elements are all working. A boiler can produce the necessary pressure and temperature but slow to get to that point, I recently had a similar situation and troubleshooting revealed that only one element out of 3 was working. Replacement of the elements cured that.
Another problem is that depending on your water quality, boilers have a tendency to build up a lot of sediment and mineral deposits, which decrease the useable volume of steam in the vessel. The only way to correct this is to remove the elements and scrape out the sediment in the boiler and replace all the elements at the same time. Its a good idea to replace the contactors too while you're at it.. This will make the unit more efficient.
Barkyboy