Any ideas where to start looking.
Mon Dec 11 2017
Error 259 - Input / Output Board #1 Communication Failure.
Wed Dec 13 2017Reply from MedWrenchMia
Wed Dec 13 2017Reply from Matt D.
Are you able to measure pressure or temperature or neither? If niether then it could be the I/O comm board 1. Check for bad connections or bad fuse 630mA @ 250V. If only one is not being measured it could be a bad sensor or bad connection to the sensor. check this site for the manual page 12-17
https://medschool.vanderbilt.edu/vbi-core-labs/files/vbi-core-labs/public_files/steris%20manual.pdf
Matt D.