Error 259 - Input / Output Board #1 Communication Failure.

Any ideas where to start looking.

   RE: Sterilizers/Autoclaves Report Post
Wed Dec 13 2017Reply from MedWrenchMia
MedWrenchMia

RPI may be able to help.



Mia DeBiase Sales & Marketing Assistant mia@medwrench.com

Report Post
Wed Dec 13 2017Reply from Matt D.
Matt D.

Are you able to measure pressure or temperature or neither? If niether then it could be the I/O comm board 1. Check for bad connections or bad fuse 630mA @ 250V. If only one is not being measured it could be a bad sensor or bad connection to the sensor. check this site for the manual page 12-17

https://medschool.vanderbilt.edu/vbi-core-labs/files/vbi-core-labs/public_files/steris%20manual.pdf