Calibration

Medrad avanta injector asking for calibration. Does anyone know how to do the calibration because I can't find it on the settings?

   RE: Contrast Media-Injector System Report Post
Wed Dec 27 2017Reply from MedWrenchMia
MedWrenchMia

ISS or Maull Biomedical may be able to help!



Mia DeBiase Sales & Marketing Assistant mia@medwrench.com

Report Post
Wed Dec 27 2017Reply from Ryan C.
avatar placemark

The calibration reminder comes up on initial boot.  This is not informing you that the systems is actually out of calibrations, but rather that the system is due according to the internal reminder.  The reminder can be changed in the settings window (as well as ensuring that the system clock/date is accurate).


Ryan Clarke - ISS  |  We Are ISS  |  888-667-1062

Report Post
Fri Dec 29 2017Reply from Eyerusalem
avatar placemark

Thank you for the information but the injector works when used manually but it is unable to retract automatically during procedures.


Report Post
Fri Dec 29 2017Reply from Ryan C.
avatar placemark

Sounds as if your systen is likely covered in contrast and in need of a thorough cleaning (part of a complete PM). 

This is commonly the contrast syringe sensor.  Remove the psi jacket and inspect the sensor pin (metallic cylinder approx 1/4" diameter).  You should be able to delress the pin and when released it returna out.  I suspect this sensor is remaining stuck.


Ryan Clarke - ISS  |  We Are ISS  |  888-667-1062

Report Post