Thu Dec 28 2017
Hiiiiiii
i need service manual of this CR machine.i have a problem of network communication. pls email- amirjemal82@gmail.com Thanks
Thu Dec 28 2017Reply from MedWrenchMia
MedWrenchMia
Thu Dec 28 2017Reply from Frank3f
This is actually a FUJI Reader  but if you et a network communication error the Reader can not talk to the Workstation Console. Check switches and network cables
Thu Dec 28 2017Reply from AMIRO
OK. we have Philips PCR-Eleva s hi-res, model - CR-IR 368
The issue is when we instell PCR softwear,there is network commnication problem and the CR is displays OFFLINE.
shear your solution with me pls
Thank you
I need service manual of FUJI CR-IR 368
Email- amirjemal82@gmail.com
take a stand alone switch and plug reader and workstation into switch making sure IP are on the same subnet and the install PCR Software, thus taking it off or the Whole network or making it stand alone while loading software
ok i will check the cables.
Thank you
I appreciate your fast response Frank, hope it will be helpful;
thank you again
Philips - PCR Eleva S Hi-res
by Philips