Thu Dec 28 2017

blurry images

I have a Siremobil compact with blurry images. Power supply voltages are spot on.

   RE: Mobile C-Arm Report Post
Fri Dec 29 2017Reply from Tomelgy
avatar placemark

When you say power supply voltages are OK, I asume you mean the image Intensifier power supply voltages!

Are the images blurred on Full Field & Zoom? If yes then the mechanical focus needs to be adjusted. If the answer is no, then you have an II power supply problem.

You can check/compare the II power supply voltages with the factory test certificate if you have one in the service manual.

Mechanical focus: First you must release the focus locking screw and then adjust focus ring.

Tom


Report Post
Fri Dec 29 2017Reply from BioLee
avatar placemark

Thanks...Correct the Intensifier supply matched up to the factory test cert.

Where is the focus locking screw and focus ring to be adjusted?


Report Post
Fri Dec 29 2017Reply from Tomelgy
avatar placemark

Is the focus blurred in both Full Field & Zoom?

Assuming you have two monitors, is the image blurred on both monitors? Do you have CRT monitors? 

Think about these things before venturing into making adjustments and making things worse.

There were two versions of focus assembly so it's not 100% easy to explain eactly.

1) Once cover off, under the Camera you see an adjustable (screw) clamping ring. Below this clamping ring you should see a 'long' Allen bolt, this is the locking bolt.

2)  Once unlocked, there is a silver ring with small horizontal holes. Use an EXACT size round bar and this ring should rotate without much resistance. You have to release radiation while adjusting.

3)  When finished retighten locking Allen bolt.

Just an FYI, the mechanical adjustment does NOT normally move unless the locking Allen bolt was loose....Or the TV Camera clamping ring is loose & the TV Camera is loose.

Tom


Report Post
Fri Dec 29 2017Reply from BioLee
avatar placemark

They have a LCD retrofit and the blur is on both displays and Full and Zoom.  

I'm thinking the unit took a good hit in transport. No damage on the outside but the system was working fine prior.


Report Post
Sat Dec 30 2017Reply from Tomelgy
avatar placemark

Yes, then I would look at the mechanical focus adjustment & camera mount/clamp.

Maybe the licking screw was loose and the focus ring has moved slightly, it does not take much!

If you have an email I can send a couple of images

Tom


Report Post
Sat Dec 30 2017Reply from BioLee
Mon Jan 13 2025Reply from Hassanbinji
avatar placemark

Hello everyone please help me out u have thesame issue 


Of having blurry images 


Report Post
Tue Jan 14 2025Reply from outinleft
avatar placemark

The first thing I always do is remove the bolts attaching the Iris/lens assy. (and camera) to the image tube, then fluoro (with a phantom in the field) until the ABS stabilizes the dose rate, then select "kv lock" to turn off the ABS, then fluoro and (carefully) lift the camera aside to see if the image on the output phosphor is sharp or fuzzy. If fuzzy = image tube problem, if sharp = camera/lens problem.


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Tue Jan 14 2025Reply from Hassanbinji
avatar placemark

Thank you for your response 

I will let you know after I tried it 


Report Post