Error Code 98

Is anyone familiar with or have a service manual for a Sakura Auto Smear CF-12DE (may also known as a cytology centrifuge, model 4325). I have a client that has one that is generating an error code 98.

   RE: Centrifuge Report Post
Mon Apr 27 2020Reply from Mirasia Beal
avatar placemark

Hello, any resolution? Same error code here...


Report Post
Tue Apr 28 2020Reply from Ross
avatar placemark

Rotational speed not reaching 50RPM within 6 seconds, or set point within 100 seconds (approx.).


Report Post
Wed Apr 29 2020Reply from Ross
avatar placemark

Sakura AutoSmear (Cytospin) - EC 98

Rotational speed not reaching 50RPM within 6 seconds, or set point within 100 seconds (approx.).

On later units, there may be a resettable circuit breaker on the rear panel. On earlier units, there is an internal fuse (FH1 - 1.6A) on a circuit board which may have blown. Replace it with an AGC 3A (312) fuse (per Sakura).


Report Post
Thu Apr 30 2020Reply from MedWrench Mandy
MedWrench Mandy

Ross shared this document.