Running a PM on my 8100 module

Running a PM on this module and when I get to the Flow Accuracy check I keep getting an error indicating (Pump) Flow blocked. I run the unit in normal mode and it infusions just fine. Any help anyone?

   RE: Infusion Pump Report Post
Fri Jan 05 2018Reply from mcwang
avatar placemark

Just had one yesterday!  this one needed a new connector, but I've had several causes in the past, including dirty AIL (I spray alcohol in there and clean w/Q-tips) Also, the pressure sensors should be around 3 volts with flow, around 1 volt w/no flow.  Go to analog sensor test and you can run it and view voltage of both sensors.  Good Luck! Mike


Report Post
Fri Jan 05 2018Reply from bobmedx
avatar placemark

All good to know! Thanks for the response. I did replace the sensors, with no fix. Replaced the logic board and got a fix. Again, thank you.

Bob


Report Post
Sat Jun 09 2018Reply from Floyd Willis
avatar placemark

This is extremely common.  All you have to do is remove the module and reattach it.  Then select yes to rerun the failed step.  I've had the pumps for years and this was recommended during our installation training.  Something about a communication error with the mainteneace software.


Report Post
Mon Jun 11 2018Reply from L Paul
avatar placemark
I agree with Floyd! very common communication issue with module in maintenance software. Open the door, do you see and hear it home. If not, Reattach, Rerun test, It should home this time. I train all new techs about this "glitch"

Report Post
Tue May 07 2019Reply from Saifa
avatar placemark

Thank you, Floyd Willis!


Knowledge is power!

Report Post