Running a PM on my 8100 module
RE: Infusion Pump Report Post

Just had one yesterday! this one needed a new connector, but I've had several causes in the past, including dirty AIL (I spray alcohol in there and clean w/Q-tips) Also, the pressure sensors should be around 3 volts with flow, around 1 volt w/no flow. Go to analog sensor test and you can run it and view voltage of both sensors. Good Luck! Mike
Report Post

All good to know! Thanks for the response. I did replace the sensors, with no fix. Replaced the logic board and got a fix. Again, thank you.
Bob
Report Post

This is extremely common. All you have to do is remove the module and reattach it. Then select yes to rerun the failed step. I've had the pumps for years and this was recommended during our installation training. Something about a communication error with the mainteneace software.