Low Na

We are having a low Na problem If we use a new bottle of dil check the problem is less but if we run a dil check on a bottle that is only 5 or 6 days old there is a shift back to low Na Is ther another way to correct this problem with out opening a new bottle every time we change the cartrige Thanks Bill carlson

   RE: Clinical Chemistry Analyzer Report Post
Mon Jan 16 2012Reply from Irfanlab
avatar placemark

This problem not because of dilution check,dont over ride the sensor, replace the sensor, check for all IMT fluids levels & expries.

Run IMT Bleeching & Dilution check, clean the sample probe & IMT probe. ( Fallow the instructions given in manual)

Irfan.


Report Post
Mon Jan 23 2012Reply from om prakash
avatar placemark

Clean monopump holes and if problem still presists then service the monopump or replace it.


Report Post
Thu Apr 19 2012Reply from sachidan
avatar placemark

Run " ringer lactate"-the the intravenous liquid for cross checking the values printed on the package itself