We are having a low Na problem
If we use a new bottle of dil check the problem is less but if we run a dil check on a bottle that is only 5 or 6 days old there is a shift back to low Na
Is ther another way to correct this problem with out opening a new bottle every time we change the cartrige
Thanks Bill carlson
Fri Oct 29 2010
Low Na
Mon Jan 16 2012Reply from Irfanlab
This problem not because of dilution check,dont over ride the sensor, replace the sensor, check for all IMT fluids levels & expries.
Run IMT Bleeching & Dilution check, clean the sample probe & IMT probe. ( Fallow the instructions given in manual)
Irfan.
Mon Jan 23 2012Reply from om prakash
Clean monopump holes and if problem still presists then service the monopump or replace it.
Thu Apr 19 2012Reply from sachidan
Run " ringer lactate"-the the intravenous liquid for cross checking the values printed on the package itself