RapidQC Complete

What is the recommended frequency for running the liquid QC for the Rapidpoint 405? And is it recommended to run all 3 levels?

   RE: Blood Gas/pH Analyzer Report Post
Thu Jan 11 2018Reply from Merdo
avatar placemark

do not stay closed for a long time. 1.2.3 All of them cause problems with congestion in current lines


Report Post
Thu Jan 11 2018Reply from Sherri Crawford
avatar placemark

I don't unstand your response "do not stay closed for a long time". Is it recommended to run the QC daily, weekly, or monthly? I've been placed in the position to maintain the analyzers with no formal training of the machine so I'm flying blind here.


Report Post
Thu Jan 11 2018Reply from Merdo
avatar placemark

run all controls daily pls


Report Post
Sun Jan 21 2018Reply from Mc Grath Carlos
avatar placemark

Manual says as you required for you qc system, its depends also how many samples you run and the values are out of ranges 