Fri Jan 12 2018

no LED on power switch, not starting

Unit has primary power, to low voltage power supply, both fuses good.

   RE: Ultrasound Systems Report Post
Tue Jan 16 2018Reply from Alexandria
avatar placemark

Which product (make and model) are you referring to?


Report Post
Wed Jan 17 2018Reply from DanVMSG
avatar placemark

Hi It is a GE LOGIQ P6 Pro, has a PST-009 power supply has line voltage to the power supply, fuses are good. None of the LEDs on the power supply or the On/Off switch light.  


Report Post
Fri Jan 19 2018Reply from dferguson
dferguson

Check if the standby 5V od APS is alive make sure AC power connector and control signal connector is plugged properly


 
 

Report Post
Fri Jan 19 2018Reply from DanVMSG
avatar placemark

Thanks, unit is down range , so will need to take a trip down, voltage detected at heat sinks inside power supply case no power supply leads will light so going towards the switching voltage.


Report Post
Fri Jan 19 2018Reply from dferguson
dferguson

Logiq A an P Series all have trouble with APS


 
 

Report Post