Sat Jan 13 2018
1043 phs control error
System will not do an exam. It will start the exam and then the table will error. I can move te table manually at the gantry no problem. All switches are good and i have 24V so the stop loop isnt open. I think possibly something with HTP is not right. Should i try encoder?
   RE: CT Scanner Report Post
Sun Jan 14 2018Reply from Chelocba
Hello.could you tell me wich model of ct? I had the same problem and I had to change the horizontal motor drive. Because the problem was in the motor encoder.
Sun Jan 14 2018Reply from Jmcmoran
It is an emotion 16 slice 2007 version.
Please. Check encoder signal.  You tried to  reload the phs  firmware in mcb? Exchange the mcb board in MAS module and then reload the firmware in booth boards.
I have swapped both MCBs and flashed them and it did not change the issue. I'm getting something from the encoder. I have loaned out my o-scope so I can't tell. The voltage changes on test points when I move the table.
Sun Jan 14 2018Reply from Ing Mario Sarduy
Hi.
I hope you are fine. Please write me to my email Mario.sarduy@gmail.com or to my cellphone +584261768710 for can know more about the system and can help you.
Best Regards