PACS configuration
RE: Mobile C-Arm Report Post
I'm assuming you don't NOT have the ISO-C 3D version!
You should have a Dicom Bridge attached, you have to conencto to Dicom Bridge using PC Anywhere software program connected to Dicom Bridge via 9pin 'D' null cable. You can connect a Monitor & keyboard directly to Dicom Bridge & make changes but its a little tricky without mouse.
To configure is not easy to explain here, there are 3 files that need to be configured with relevant details, detail that the IT person would have provided like;
Local IP
Local AET
Destination IP
Desrination AET
Destination Port number
There are two versiosn of Dicom Bridge, Windows 2000 & later Windows XP.
Report Post
Thanks very much for the info Tomelgy.
This site is approx. 1.5 hours away and I’m trying to do prep work. Do you believe I may find these instructions in the Siemens service documentation?
Thanks again!
Report Post
Hopefully, if the customer keeps them safe. or they are inside the monitor trolley.
Has the system been configured connected to PACS before?
I ask because the Dicom Bridge is an Option usually purchased at time of order!
Regards
Tom
Report Post
Thanks for your help Tom.
The system has been connected to PACs before.
Our health system is changing PACs and worklist servers.
Randy
Report Post
That will make things a little easier, you will be able to compare existing configuration with new data.
Just suggestion. Make copies on the existing files, copy & rename ***.old for example just in case things don'y go well.
Tom