flow cell adjust

how do you adjust the flow cell for the differantials in a Cell Dyn 3200

   RE: Hematology Analyzer Report Post
Tue Jan 16 2018Reply from MedWrenchManager
MedWrenchManager

MYCO may be able to help with this question. 


Report Post
Tue Jan 16 2018Reply from Wytec Lda
Wytec Lda

you will require a service manual and some special tools for this assignment ,do you have these things ?


PRangel

Report Post
Tue Jan 16 2018Reply from LabDr
avatar placemark

yes i have the service manual and the latex particles. i thought i had it alighned good, the CV's are where they need to be. controls pass. but the patient samples are all running a little high on the Baso. i did replace the laser. could that be the difereance from a week laser to a strong laser. Thank you.


Report Post
Tue Jan 16 2018Reply from Wytec Lda
Wytec Lda

The Baso can not be higher regardless of having changed the laser tube .

make a fresh latex ( latron ) and fine tune the CV's to close o perfect 

Did you change the laser and laser Power supply unit ? , some times it is better to change the power supply as well because a new tube could wear out the older laser power supply .

With good alighment 

good CV´s

I feel that you are nearly there . 

Are you using fresh blood samples for patient samples , from how many hours ago ? Are the baso results high on all the patient samples ? If possible draw blood and run a fresh whole blood sample .

baso raise as whole blood in the tube gets older . 


PRangel

Report Post
Tue Jan 16 2018Reply from LabDr
avatar placemark

i would lik to send you pictures of the CV's i got, patient sample, and controls. to look at the histogram and tell me what you think. i could email or text them. i dont think there is a way to do it on this site.


Report Post
Tue Jan 16 2018Reply from Wytec Lda
Tue Jan 16 2018Reply from MedWrenchManager
MedWrenchManager

@labDr-- there is also the opton to add images here. just click on the on the image icon when posting your response.


Report Post
Tue Jan 16 2018Reply from marcosanti2003@yahoo.es
avatar placemark

The following items are necessary:
woc fine procedure.
Excel spreadsheet to find new gains.
laser power meter.
latex of 7 um.

laser protection glasses 
fresh blood from patient with less than 04 hours.

the information and procedures I can send it to you by mail


Report Post