5810R with power interruption

Hello I have a 5810R and when use temperature control lowering below 12C tripped power and turned it off. So it look like when compressor works draw a lot of current and tripped. I have a spare 5810R I can replace some parts. Which part should look at it first. Thank you

   RE: Centrifuge Report Post
Fri Jan 26 2018Reply from Jeff M
avatar placemark

Does the power trip as soon as the compressor starts running, or after a while? Does the compressor sound normal? Does it run ok if you set the refrigeration at room temperature or higher? Does the power trip if you set a lower temperature and close the lid without starting a spin?

Can you try a different power receptacle, not on the same branch as the one the machine is plugged into now?

If the power fails as soon as the compressor starts, and if it happens the same way on another receptacle, try the compressor start-up components first, if it has any and if they are accessible.