9900 overshooting KV

Hi I have a system that has no errors being reported but while testing I found the KV in some shots( mostly higher and max dose shots, ) are over 6% of kv on c arm and monitor from the reading on my meter. Some are as much as 15% higher on the meter. Ive tried a different meter and get the same results. Ive done filament cals and saved to system twice. The complete generator cal has been done . Ive tried a new high voltage reg pcb

   RE: Mobile C-Arm Report Post
Wed Feb 14 2018Reply from Rapidxray.biz
avatar placemark

Have you scoped the KV and MA waveforms to see what they are doing and check for tube spits? Any specific MAS settings? Film mode, Digital spot mode, fluoro mode, or high level fluoro? When you say SHOTS I am assuming radiographic but not fluoro..... 