Help ... Manual service DRS 601
RE: Slide Stainer/Prep Report Post
We have 220 V and we need to change the polarity from 110 V to 220 V
for that reason we need the service manual.
with regards
With Regards,
Capsimpex sarl
http://capsimpexsarl.webnode.fr
It is section 8-7 in the service manual. For 220/240V operation, remove the short on the right-hand end of the terminal block (on the 24V switching regulator).
[(Note to administrator: I have a copy I can send in. Tell me how.]
Dear Ross
thank your for your help, can u send it through my email adress at :
with regards
With Regards,
Capsimpex sarl
If you would like to share with the MedWrench community, please email your document to info@medwrench.com and we will upload it to the product page on your behalf.
Dear ross
We resolve the change as you have explained to me. We remove the short . Now it run on 220V.
Thank you for your help.
With Regards,
Capsimpex sarl