Valleylab shuts off soon after its turned on.
Hi , the power switch is a common problem when the device is resetting alon.
You can check this, pushing slowly the botton,and you'd hear a false contact
Sometimes this issues could cause fall in CPU board.
In fxs made after 2013 you should change the switch every year
Rodrigo
As the instructor for all Valleylab generators, I must say I've never seen either of those error codes in the 8+ years I taught the Force FX-C. I would first try to swap out the low voltage power supply to see if that makes a difference (it's only three connectors and four screws). After that, I would swap the controller board to see if that makes a difference.
I hope that works for you, but if you have any questions, feel free to contact me at www.biomedsupport.com or 303-525-3393.
Dale Munson
Dale Munson