Cleaning

What is recommended for cleaning the inside of this equipment? Need for joint commission standards.

   RE: Centrifuge Report Post
Thu Feb 22 2018Reply from Bob S
avatar placemark

Clean bowl area periodically. Remove aluminum shields, unscrew black and white knob counterclockwise, and carefully lift the rotor off the motor shaft. Wipe the interior area of the bowl with a cloth dampened with a suitable disinfectant.

Clean rotor with a solution of sodium hypochlorite in concentrations (of sodium hypochlorite) not to exceed 5%. Rinse with warm water then dry rotor. CAUTION: DO NOT AUTOCLAVE ROTOR.

Check rotor periodically for structural integrity. REPLACE ROTOR IF WORN OR CRACKS ARE VISIBLE. Replace with V7000 aluminum rotor only.