Wed Feb 28 2018
99 degree shown on the digit of oven
I have a vip with a problem of the heat of the oven and the rotor. When i switch on . In manual mode the temperature shown is 99 degree but the oven and rotor still cold. I let it for one hour then i listen a bip when i check are the error list E34 E36. I checked the triac are perfect. Doas anyone had this problem before. Thanks for all
   RE: Tissue Processor Report Post
Wed Feb 28 2018Reply from Paul Moyer
The Thermistor is shorted. Failure of the parafin oven temp.
Fri Mar 02 2018Reply from CAPSIMPEX Morocco
CAPSIMPEX Morocco
Ok Paul
i will check the thre thermistor and will let you know .part number :PB-41E . thank you
With Regards,
Capsimpex sarl
http://capsimpexsarl.webnode.fr
Tue Mar 13 2018Reply from CAPSIMPEX Morocco
CAPSIMPEX Morocco
Dear Paul
I Checked the three thermistor and they are good.
maybe the photocopler are damaged or IC2 circuit what do u think?
thanks
With Regards,
Capsimpex sarl
Wed Mar 14 2018Reply from Paul Moyer
I do not have schematics handy but the optocoupler would be a good place to start. If you have one I would replace it just make sure it is the same type.
Wed Mar 14 2018Reply from CAPSIMPEX Morocco
CAPSIMPEX Morocco
I change the optocoupler Ic2 named on the pc board and now its ok. The machine work correctly i have now the error E 13 , i will check now  whats the problem
With Regards,
Capsimpex sarl