When I turn on the unit it display M010 Build-in self test fail. Also during expuser, sometimes apper an error M208.
RE: Mobile C-Arm Report Post
Tue Mar 13 2018Reply from Alex Wright

Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
Tue Mar 13 2018Reply from chris?

Problem is the succo bd inside the c-arm is intermittent will have to replace it the sram battery on the bd is low causing the errors intermittent codes
Report Post
Tue Mar 13 2018Reply from bfainsztein
Wed Oct 27 2021Reply from juventino
Wed Nov 03 2021Reply from semmedical
Thu Nov 04 2021Reply from Ben C

Although it is your choice to go offline with your discussion, we are sure that other MedWrench users would appreciate it if you kept your discussion within the thread. This will help to keep a permanent record of the issue you are facing and how it was solved to help others in the community that may be facing a similar challenge.