Permanent Lines on Xray Images
RE: Computed Radiography Report Post

Usually if the lines run the direction of the plate travel and cover the entire length of the plate its caused by dust blocking the laser. Have you used the manual cleaner on the side of Prima T scanner?
Report Post

I have not any experience with Prima T2, but lots of it with Capsula, FCR5000 and XG1.
How do these lines look like? check with a 35x43 plate. Line parallel to the 43cm (in the direction of transport) from top to bottom can mean dirt on the mirror close to the light-collector in the subscanner. Another cause might be a long blond hair disturbing the laserbeam...
Henk1951
Report Post

Is this problem solved???
lines - in the ip transport direction - are almost always caused by dirt in the subscanning unit. Dirt is blocking the light transport from the ip to the PMT. E.g the edge of the light collector or the rotatable mirror in front of this edge. Start with using the cleaning button on the CL. When this is not bringing the solution start cleaning the the mirror and the lightcollector edge, all of coure with much care. Another cause might be a hair in front of the output windo of the laser unit.
Henk1951