Thu Mar 22 2018

open door alarm

we have a different problem, the open door alarm won't stop even when the door are closed. it's a sorvall RC6 any advice? Tanks.

   RE: Centrifuge Report Post
Thu Mar 22 2018Reply from Alex Wright
Alex Wright

Hello, try contacting Ozark for assistance 



Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com

Report Post
Thu Mar 22 2018Reply from DFMK6
avatar placemark

sounds like a door lock solenoid or sensor issue. I'd start there.


Report Post
Thu Mar 22 2018Reply from sor
avatar placemark

yes this is what i suspect,  do you now where is sensor exactly, and there is one or more?


Report Post
Thu Mar 22 2018Reply from DFMK6
avatar placemark

Check to make sure you are getting 5V locked and 0V open with your DMM from the left and right door lock solenoids.

The service manual is located under the documents tab if you need help locating them.


Report Post
Fri Mar 23 2018Reply from sor
avatar placemark

okk tank's i will do it,

and how i can Check program


Report Post
Wed Mar 28 2018Reply from david.rodriguez@bcw.edu
avatar placemark

I've had similar issues with our RC-3BP Plus and the latch was broken.  I think theres a bad run of latches