The problem with DR detector

Hello! I have the problem with DRX-1 wireless detector. Calibration without x-ray fails... I think the detector have some area of bad sectors. Can i solve the problem through the calibration or other method? Thank you!

   RE: Digital Detectors Report Post
Fri Mar 23 2018Reply from Alex Wright
Alex Wright

See if MIT can help on this one!



Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com

Report Post
Mon Mar 26 2018Reply from SWH
avatar placemark

First dis-associate the detector from the DRX-1 system. As though you are going to use another dectector in place of the one normally associated with that system. The re-associated it with the system and perform the calibrations. This can be done with out access to the service software.


Report Post
Mon Mar 26 2018Reply from Sergio123
avatar placemark

Thanks for the answer. In order to remove the detector from the system, the master operator password is required. Just turning off the detector in the program did not help.


Report Post
Mon Mar 26 2018Reply from SWH
avatar placemark

To access the Key Operator software, should be KeyOp, 1313, maybe keyop01, 1313.


Report Post
Mon Mar 26 2018Reply from Sergio123
avatar placemark

Thank you very much! tomorrow I will definitely check this password and report the results


Report Post
Tue Jan 07 2020Reply from cronos1999
avatar placemark

Could You fix the problem?