I have the problem with DRX-1 wireless detector. Calibration without x-ray fails...
I think the detector have some area of bad sectors. Can i solve the problem through the calibration or other method?
Thank you!
Fri Mar 23 2018
The problem with DR detector
Fri Mar 23 2018Reply from Alex Wright

See if MIT can help on this one!
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
Mon Mar 26 2018Reply from SWH

First dis-associate the detector from the DRX-1 system. As though you are going to use another dectector in place of the one normally associated with that system. The re-associated it with the system and perform the calibrations. This can be done with out access to the service software.
Report Post
Mon Mar 26 2018Reply from Sergio123

Thanks for the answer. In order to remove the detector from the system, the master operator password is required. Just turning off the detector in the program did not help.
Report Post
Mon Mar 26 2018Reply from SWH
Mon Mar 26 2018Reply from Sergio123

Thank you very much! tomorrow I will definitely check this password and report the results