Error 384 battery end of life
RE: Mobile C-Arm Report Post

Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post

send me an email address to send the manual to. 384 normally means the Energy Storage Unit (ESU) the batteries need to be replaced. it is a timed event and to reset it you need to access the BVScope software on the system and select replace ESU. Philips being Philips does not give CSIP level 0 access to this function.
Report Post

Help
the batteries need to be replaced. it is a timed event and to reset it you need to access the BVScope software on the system and select replace ESU. Philips being Philips does not give CSIP level 0 access to this function.
what??
Report Post

Although it is your choice to go offline with your discussion, we are sure that other MedWrench users would appreciate it if you kept your discussion within the thread. This will help to keep a permanent record of the issue you are facing and how it was solved to help others in the community that may be facing a similar challenge.