Thu Apr 05 2018

Symbia S error code

symbia s error code 1201-0057

   RE: Nuclear Medicine-Gamma Camera Report Post
Thu Apr 05 2018Reply from Alex Wright
Alex Wright

MedWrench is available to help people with specific questions and troubleshooting problems. Could you please be more specific with the issue that you are having with your instrument? The MedWrench team will do their best to assist you in solving the issue. This will also help other community members who face similar challenges.



Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com

Report Post
Thu Apr 05 2018Reply from Tom Roberts
avatar placemark

I figured it out.

The retraction limit switch for the auto Q/C line source was not engaging when the line source got to its mechanical end limit. The drive motor would just keep spinning until a lack of motion reported from the encoder timed out and generated the error. I heard a ratcheting type noise caused by the motor drive gear slipping over the teeth of the belt drive gear after the line source went as far as it could but before the motor stopped spinning.

I removed the auto Q/C source shelf and moved the end limit switch so that the switch would engage when the line source got to the fully retracted position.

The scanner was fully functional after that.


Report Post
Thu Apr 05 2018Reply from ISStech3
avatar placemark

@ Tom, great thanks! My co-workers are working on it and will pass this along.


Report Post