Blank Screen

Unit suddenly stopped working just about to shoot an x-ray. Charge light is full, can drive the unit but nothing else. Push the button to start and it seems to, but the windows screen never comes on and it shuts itself down. Light still on the push buttons to select parts, collimator light works by pushing the light button. Does not display any error codes to know what is wrong. What could this be?

   RE: Mobile Radiography Report Post
Sun Apr 08 2018Reply from Arthur
Arthur

It sounds like you do not have continuous power to your Windows PC. There is a power supply on the left side which has connectors feeding power to the PC and it loses good connection. Usually I just have to remove the connector and clean the pins and then it works.

The power supply in on the right side near the operator drive position. It is about 3 inches wide and 8 inches long. If you send an email address I will send you a picture from the parts manual, page 46 item # 151

 

Art


Report Post
Sun Apr 08 2018Reply from Arthur
Arthur

On medwrench there is a parts manual for the Evolution. In that manual on Page #49 item # 18 will give you a general idea of where the power supply is on the Dart Model.

 

Art


Report Post
Sun Apr 08 2018Reply from drruc
Wed Apr 11 2018Reply from drruc
avatar placemark

The Relay HG2-AC100V-F AP6124F seems to be the culprit here. Where can I find this relay to buy? thanks


Report Post
Thu Apr 19 2018Reply from drruc
avatar placemark

Pls, I still need help. I Changed the relay and still will not turn on. I then noticed the computer gets power. There is a power button on the front of it that if you push it comes on. I then took the computer out and changed the button battery but it still will not come on automatically as before. I now have the back of the unit open. What do I do to the get this working? thanks


Report Post
Thu Apr 19 2018Reply from Arthur
Arthur

The computer button either sets the computer to ON or OFF. It the switch is in the OFF mode it will not come on when you push the blue computer start button on the top panel. Push the small computer ON/OFF button "Once" and the computer will start. Then power down the system and try to power it on again from the Blue start button on the top.

 

Art


Report Post
Tue Dec 03 2019Reply from abharris
avatar placemark

We're having this issue too.  When I remove the panel on the right side, one of the HDSP A101 LED display is blinking red.  Could this be our issue?


Report Post
Thu Sep 08 2022Reply from Ahmed S. Ali
avatar placemark

 I have same your problem the blank monitor but i didn't find that relay HG2-AC100V-F AP6124F in spare parts manual

where did you find it please

 


Report Post