Water in sample tube

Our BS-200 delivers water in sample tubes, while running the analysis. Where could be the problem? We called a maintenance engineer, but as usual in our country he will not be available before few days.

   RE: Clinical Chemistry Analyzer Report Post
Fri Apr 13 2018Reply from Alex Wright
Alex Wright

Hello, Have you contacted MYCO for assistance? Were they able to helo with this?



Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com

Report Post
Wed Apr 18 2018Reply from Joao Ambrosio
avatar placemark

Check if the probe is dripping. If it is, probably the Sample Valve (BA30-21-06469) is defective.

 


Joao Ambrosio
Report Post
Wed Apr 18 2018Reply from Joao Ambrosio
avatar placemark

You can also check if the probe is not obstructed and if the sample tubing is not punctured.


Joao Ambrosio
Report Post
Fri Apr 20 2018Reply from Samir Sahlab
avatar placemark

Many thanks for your kind replies. The problem was coagulum on pump membrane. This is what maintenance engeneer stated, Thanks.