Mindray BS-380 - Error C0714
RE: Clinical Chemistry Analyzer Report Post
Hello! Have you tried checking to see if MYCO can help with this? That is what we would recommend!
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
friend i had this same problem last month my email is jad738@hotmail.com .... look!!!!!! the first thing you must do its upgrade all control softwares thats depend if the mother board its old you can use the upgrade software 1.8 because the new model you can upgrade until 1.9 or 1.10.
**Version Information**
Unit Name Version 1.9F
--------------------------------------
Main Unit: 1.9
SampMixer Unit: 1.3
Reag2Probe Unit: 1.6
SampProbe Unit: 1.7
Clean Unit: 1.6
ReagTray Unit: 1.4
SampTray Unit:1.3
ReacTray Unit:1.2
Temperatureunit Unit:1.1
**Upgrade Notice**
1. Please DO remember backup all parameters.
2. Please follow the upgrading guidance to upgrade the control software
but you must do it step by step from the minor version to the highest
IF YOU WHAT I TALK YOU THIS IS THE FIRST STEP
BS-380&390 Control Software Upgrading Guidance
Note:
If upgrading from V1.8 and below version to V1.9 or above, there are special steps to be performed,
otherwise the upgrade can’t be completed. Please pay attention to the page 6 “The Special procedure for
upgrading from V1.8 and below version to V1.9 or above ".
Upgrade procedure:
1. Turn off the machine. Make sure the main board jumper is on Pin2&3 where disables the parameter
modification.
2. Backup the units parameters
1) Open the main switch and analyze switch;
2) Open the computer, then run the debug software;
3) Go to parameter configuration interface, save all the units parameters as txt file.
4) Exit the debug software, turn off the analyze switch.
NoteThis step is very important!
3. Run “Upgrade.exe” in the control software folder.
4. Turn on both main switch and analyzing unit switch.
Note: Machine should be powered on AFTER running the file “Upgrade.exe”. Reversing the sequence
will cause communication failure.
5. 1) Choose “COM1” from the drop down list.
2) Choose “MainUnit” from the drop down list.
3) Click the browse button then choose file “MainUnit.bin” from the control software folder.
4) Click “Download” to start upgrading main unit.
WHEN YOU FINISH IT YOU MUST CLEAR ERROR MESSAGE IN THE DEBBUG SOFTWARE
IF YOU NEED MORE INFORMATION EMAIL ME
Report Post
mAYBE if you change the photometric values i hope you saved that first & after you modify that in main unit and after in adquire ????????????????? but the problem is when you change those values and the lamp shiner you look good values in the dark current but you open the operator software appears the message fails : THE DARK CURRENT IS TOO HIGH
Report Post
Hello, any person can helpme to download BS380 Debug software, thanks my email is crisferquezada339@yahoo.com
Report Post
I have the same problem, lamp new, looks OK everithing , values OK but still the message lamp weak. How did you solve it?
Thank you
Report Post
I can t understand how the threshold of the lamp weak is set in the operating software....
In the service manual they say this threshold is set based of the 340 nm parameter gain, but I don t understant how ...
This value can only be read in the operating software and can t be modified.
Could anyone explain me, may be I can adjust some parameters if I could understant the reason behind the threshold